08-28-2014, 06:18 PM
Hi everyone. I have a Digital Aquatics reef keeper with SL2 module on it. I calibrate probe with 53.0 mS conductivity solution then i used solution to calibrate my Vertex refractometer to 35.0 ppt, this is crossover conversion. So now both are calibrated to the same level. DA now says 45.5mS,29.5ppt(corals looking sluggish thats why i recalibrated) and then Vertex says 60mS,40.5ppt. How can they be calibrated to same then one test high and one test low. Any ideas out there? im wondering if i need to replace probe.
53mS=35.0ppt=1.0259SG
Scott
53mS=35.0ppt=1.0259SG